  • Select the area you would like to search.
  • ACTIVE INVESTIGATIONS Search for current projects using the investigator's name, institution, or keywords.
  • EXPERTS KNOWLEDGE BASE Enter keywords to search a list of questions and answers received and processed by the ADNI team.
  • ADNI PDFS Search any ADNI publication pdf by author, keyword, or PMID. Use an asterisk only to view all pdfs.

Question

Question Posted 02/28/18:
We recently downloaded a batch of ADNI 3 preprocessed MPRAGE images. I noticed that a couple subjects scanned on the Siemens Verio had two images on the IDA, one with MT1;N3m and another image with MT1;N3m;GradWarp processing. The image with the gradwarp processing has a "MPRAGE_ND" in the description. Is there some reason that the Verio scans require the gradwarp? Is the 3D gradient warping correction not applied by this particular scanner? If we wish to use the 3d gradient warp corrected scan should we use the MPRAGE_ND scan? Thank you very much
Response posted 03/01/18 by ADNI MRI:
Thanks for your question to the ADNI MRI Core.

The Siemens Verio system provide distortion corrected and Non distortion corrected (MPRAGE_ND). We would recommend that you only use the standard scan and not use the ?ND? unless you want non grad warp corrected scans.

Thanks

ADNI MRI
Go back to list of topics >